Dear All,
I have had an enquiry from someone from another Department who has an artefact affecting their imaging which looks like a diffraction effect (e.g. Newton rings) when doing transmitted light microscopy. I had a look at their microscope and camera system and even when Koehler illumination is set up properly, you still get this effect. The microscope is a rather old Nikon inverted microscope with a Coolpix attached. The objectives, (phase contrast), are not infinity corrected and I'm assuming that there is also a problem with the tube length for the camera. The rings are there irrespective whether specimens are on glass or in plastic dishes. I don't think much can be done optically (correct me if I'm wrong) but I wondered if some image processing could help. My ImageJ question is this: Is it possible to use some kind of image processing (e.g. image subtraction) to remove this artifact? I have tried a few things out (e.g. Image calculator, rolling ball) without much success so far. I have posted 2 images at the following address: http://www.health.auckland.ac.nz/biru/exptal_images/index.html There are also links to larger images although the original images are even larger than these. Please note that I do know that these images are appalling! I asked one of the students to send me one image with tissue (glass slide) and one without at the same focal plane and this is what I got! I don't think they set up the microscope very well. However, you can see the rings clearly on both images so I hope they will inspire someone to suggest a solution. I would be very pleased to hear of any potential solutions. Cheers, Jacqui. Jacqueline Ross Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland, NEW ZEALAND Tel: 64 9 373 7599 Ext 87438 Fax: 64 9 373 7484 http://www.health.auckland.ac.nz/biru/ <http://www.health.auckland.ac.nz/biru/> |
Jacqui,
This is, unfortunately, an optical artifact of the coolpix cameras. We have seen this primarily with the Coolpix 995, but I understand that it occurs in others as well. Our solution is to take a background image and subtract it from the image of the sample. However, this is a particularly tricky thing to do with color images. Since we see this problem most often with DIC images, we just convert to greyscale and work with that. If color is important, you might consider separating the images into R,G,B images, processing each of them, and then reassembling with RGB merge, or take a look at the Color Balance program by Gabriel Landini. (http://www.dentistry.bham.ac.uk/landinig/software/software.html) I haven't used it, but the design does make sense. The best way to eliminate the artifact, I'm afraid, is to get a different camera. Joel Date sent: Tue, 18 Jul 2006 18:40:55 +1200 Send reply to: ImageJ Interest Group <[hidden email]> From: Jacqui Ross <[hidden email]> Subject: Optical artefact removal To: [hidden email] > Dear All, > > > > I have had an enquiry from someone from another Department who has an > artefact affecting their imaging which looks like a diffraction effect > (e.g. Newton rings) when doing transmitted light microscopy. > > > > I had a look at their microscope and camera system and even when > Koehler illumination is set up properly, you still get this effect. > The microscope is a rather old Nikon inverted microscope with a > Coolpix attached. The objectives, (phase contrast), are not infinity > corrected and I'm assuming that there is also a problem with the tube > length for the camera. The rings are there irrespective whether > specimens are on glass or in plastic dishes. I don't think much can be > done optically (correct me if I'm wrong) but I wondered if some image > processing could help. > > > > My ImageJ question is this: > > > > Is it possible to use some kind of image processing (e.g. image > subtraction) to remove this artifact? I have tried a few things out > (e.g. Image calculator, rolling ball) without much success so far. > > > > I have posted 2 images at the following address: > http://www.health.auckland.ac.nz/biru/exptal_images/index.html There > are also links to larger images although the original images are even > larger than these. > > > > Please note that I do know that these images are appalling! I asked > one of the students to send me one image with tissue (glass slide) and > one without at the same focal plane and this is what I got! I don't > think they set up the microscope very well. However, you can see the > rings clearly on both images so I hope they will inspire someone to > suggest a solution. > > > > I would be very pleased to hear of any potential solutions. > > > > Cheers, > > > > Jacqui. > > > > Jacqueline Ross > Biomedical Imaging Research Unit > School of Medical Sciences > Faculty of Medical & Health Sciences > The University of Auckland > Private Bag 92019 > Auckland, NEW ZEALAND > > Tel: 64 9 373 7599 Ext 87438 > Fax: 64 9 373 7484 > > http://www.health.auckland.ac.nz/biru/ > <http://www.health.auckland.ac.nz/biru/> > > Joel B. Sheffield, Ph.D Department of Biology Temple University Philadelphia, PA 19122 Voice: 215 204 8839 e-mail: [hidden email] |
In reply to this post by Jacqueline Ross
Hi
Hi Jacqueline, Looking at your image, I think you have sort of a light leek somewhere in your optical . Do you have 2 light sources on your microscope? Or one port that is not light proof? I had that problem with a fluorescence Nikon upright microscope when we where acquiring images in transmitted light and that the mercury lamp was on (at that time we where missing an empty filter cube in the filter turret) but we could avoid the diffraction effect by closing the manual fluorescence shutter on the microscope. But for your actual images, I guess you could easily do a shading correction by: 1- getting a background image (a spot on your slide but out of the specimen area) with the same parameters of exposure time and binning than your specimen image 2- than you divide your specimen image by the background image to get your shading corrected image. Monique Vasseur Microscopie et imagerie Département de biochimie Université de Montréal tél. (514) 343-6111 poste 5148 -----Message d'origine----- De : ImageJ Interest Group [mailto:[hidden email]] De la part de Jacqui Ross Envoyé : 18 juillet 2006 02:41 À : [hidden email] Objet : Optical artefact removal Dear All, I have had an enquiry from someone from another Department who has an artefact affecting their imaging which looks like a diffraction effect (e.g. Newton rings) when doing transmitted light microscopy. I had a look at their microscope and camera system and even when Koehler illumination is set up properly, you still get this effect. The microscope is a rather old Nikon inverted microscope with a Coolpix attached. The objectives, (phase contrast), are not infinity corrected and I'm assuming that there is also a problem with the tube length for the camera. The rings are there irrespective whether specimens are on glass or in plastic dishes. I don't think much can be done optically (correct me if I'm wrong) but I wondered if some image processing could help. My ImageJ question is this: Is it possible to use some kind of image processing (e.g. image subtraction) to remove this artifact? I have tried a few things out (e.g. Image calculator, rolling ball) without much success so far. I have posted 2 images at the following address: http://www.health.auckland.ac.nz/biru/exptal_images/index.html There are also links to larger images although the original images are even larger than these. Please note that I do know that these images are appalling! I asked one of the students to send me one image with tissue (glass slide) and one without at the same focal plane and this is what I got! I don't think they set up the microscope very well. However, you can see the rings clearly on both images so I hope they will inspire someone to suggest a solution. I would be very pleased to hear of any potential solutions. Cheers, Jacqui. Jacqueline Ross Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland, NEW ZEALAND Tel: 64 9 373 7599 Ext 87438 Fax: 64 9 373 7484 http://www.health.auckland.ac.nz/biru/ <http://www.health.auckland.ac.nz/biru/> |
Hi
It is right that this is a specific coolpix artefact, it is present for models after CP-995 (4500, 5000) It is due to the photosites size (usually for digital cameras the improvements for general photography are bad things for microscope microphotography). Well, you can diminish this artifact in manual mode trying different aperture settings (f/2.8, f/5,6, etc.) and of course adjust the right exposure with the speed. After this you will need also to play with the microscope aperture diaphragm (in the condenser) in order to optimize results. I hope it helps for your future picture, for old pictures maybe somme band pass filter correction ? Leon Le 18 juil. 06 à 15:22, Vasseur Monique a écrit : > Hi > > Hi Jacqueline, > > Looking at your image, I think you have sort of a light leek > somewhere in your optical . Do you have 2 light sources on your > microscope? Or one port that is not light proof? I had that > problem with a fluorescence Nikon upright microscope when we where > acquiring images in transmitted light and that the mercury lamp was > on (at that time we where missing an empty filter cube in the > filter turret) but we could avoid the diffraction effect by closing > the manual fluorescence shutter on the microscope. > > But for your actual images, I guess you could easily do a shading > correction by: > > 1- getting a background image (a spot on your slide but out of the > specimen area) with the same parameters of exposure time and > binning than your specimen image > 2- than you divide your specimen image by the background image to > get your shading corrected image. > > Monique Vasseur > Microscopie et imagerie > Département de biochimie > Université de Montréal > tél. (514) 343-6111 poste 5148 > -----Message d'origine----- > De : ImageJ Interest Group [mailto:[hidden email]] De la part > de Jacqui Ross > Envoyé : 18 juillet 2006 02:41 > À : [hidden email] > Objet : Optical artefact removal > > Dear All, > > > > I have had an enquiry from someone from another Department who has an > artefact affecting their imaging which looks like a diffraction effect > (e.g. Newton rings) when doing transmitted light microscopy. > > > > I had a look at their microscope and camera system and even when > Koehler > illumination is set up properly, you still get this effect. The > microscope is a rather old Nikon inverted microscope with a Coolpix > attached. The objectives, (phase contrast), are not infinity corrected > and I'm assuming that there is also a problem with the tube length for > the camera. The rings are there irrespective whether specimens are on > glass or in plastic dishes. I don't think much can be done optically > (correct me if I'm wrong) but I wondered if some image processing > could > help. > > > > My ImageJ question is this: > > > > Is it possible to use some kind of image processing (e.g. image > subtraction) to remove this artifact? I have tried a few things out > (e.g. Image calculator, rolling ball) without much success so far. > > > > I have posted 2 images at the following address: > http://www.health.auckland.ac.nz/biru/exptal_images/index.html > There are > also links to larger images although the original images are even > larger > than these. > > > > Please note that I do know that these images are appalling! I asked > one > of the students to send me one image with tissue (glass slide) and one > without at the same focal plane and this is what I got! I don't think > they set up the microscope very well. However, you can see the rings > clearly on both images so I hope they will inspire someone to > suggest a > solution. > > > > I would be very pleased to hear of any potential solutions. > > > > Cheers, > > > > Jacqui. > > > > Jacqueline Ross > Biomedical Imaging Research Unit > School of Medical Sciences > Faculty of Medical & Health Sciences > The University of Auckland > Private Bag 92019 > Auckland, NEW ZEALAND > > Tel: 64 9 373 7599 Ext 87438 > Fax: 64 9 373 7484 > > http://www.health.auckland.ac.nz/biru/ > <http://www.health.auckland.ac.nz/biru/> > > > Leon Espinosa [hidden email] Laboratoire des Rickettsies du Pr. RAOULT UMR CNRS 6020 Fac. de Medecine de la Timone 27 Bd Jean Moulin 13005 Marseille tel 04 91 38 55 17 fax 04 91 38 77 72 portable 06 79 25 97 40 |
In reply to this post by Joel Sheffield
On Tuesday 18 July 2006 14:15, Joel Sheffield wrote:
> them, and then reassembling with RGB merge, or take a look at the > Color Balance program by Gabriel Landini. > (http://www.dentistry.bham.ac.uk/landinig/software/software.html) The plugin Joel mentioned (I guess that you mean Colour_Correct) unfortunately does not correct this kind of artifact. I am not sure if the following (ratio) works in phase contrast (it works on bright field, though): Capture a shot with the light path off (all black), Call it Darkfield (this will compensate 'hot' pixels). Capture a shot of the light path on (no specimen, all background) call it Brightfield (this will compensate uneven background). Put the specimen in and capture; call it Specimen Using the image calculator do: Subtract Brightfield-Darkfield, call it Divisor Subtract Specimen-Darkfield, call it Numerator Using Calculator_Plus plugin (in IJ site) do: Divide Numerator by Divisor and multiply by 255. The result should have a flat background and white will appear white as far as you do not change at all the settings of the camera between shots (time, exposure, autobalance etc) or the microscope setting (light, condenser, magnification). One can improve the S/N ratio by taking average shots instead of single ones (the improving factor is, I think, the square root of the number of shots). The Divisor image can be used for subsequent shots if the light source is stable and you do not change magnification. The Darkfield image can be used as well if you do not change camera settings. Some cameras have auto exposure which makes this kind of procedure impossible to apply. Cheers, Gabriel |
In reply to this post by Leon Espinosa
Hi all,
>It is right that this is a specific coolpix artefact, it is present >for models after CP-995 (4500, 5000) It is due to the photosites size >(usually for digital cameras the improvements for general >photography are bad things for microscope microphotography). Hmm, are you referring to some kind of moire effect? However, I could not understand how this should show up on a presumed homogeneous (resp. non-periodic) background!? As I understand they are using non-infinity objectives in some kind of infinity position, I would assume (the order of this effect depends greatly on the focal length of the objctive used) that they need to defocus so much that a normally unnoticed (becaused totally out of focus) internal reflex becomes near focus and noticeable, maybe this is even a bad coincidence with the reflection from the "cover glass" (or whatever, I don´t know the layout of coolpix cameras) in front of the chip, so it might disappear with other cameras, because they have that glass in another position! Joachim ______________________________________________________________________ This email has been scanned by the MessageLabs Email Security System. For more information please visit http://www.messagelabs.com/email ______________________________________________________________________ |
It is not at all clear where the artifact comes from. We have had
the same effect with infinity corrected lenses. The size of the rings is independent of the objective magnification. It is unaffected by microscope focus. I like Joachim's suggestions about the structure of the camera. It could easily be an internal reflection problem of some sort. It does not appear to be a function of the camera focal length --at least within the zoom ranges that are used for microphotography. If the sample is complex, and fills the field, i.e. a tissue section, the effect of the artifact is minimal. However, if there is an extensive area of background, or if the illumination is dim, the artifact becomes prominent. Interestingly, it is sufficiently consistent that it can be removed by background subtraction procedures that use a "no sample" image, as long as intensity is kept constant. Joel > Hi all, > > >It is right that this is a specific coolpix artefact, it is present > >for models after CP-995 (4500, 5000) It is due to the photosites size > >(usually for digital cameras the improvements for general > >photography are bad things for microscope microphotography). > > Hmm, are you referring to some kind of moire effect? However, I could > not understand how this should show up on a presumed homogeneous > (resp. non-periodic) background!? > > As I understand they are using non-infinity objectives in some kind of > infinity position, I would assume (the order of this effect depends > greatly on the focal length of the objctive used) that they need to > defocus so much that a normally unnoticed (becaused totally out of > focus) internal reflex becomes near focus and noticeable, maybe this > is even a bad coincidence with the reflection from the "cover glass" > (or whatever, I don´t know the layout of coolpix cameras) in front of > the chip, so it might disappear with other cameras, because they have > that glass in another position! > > Joachim > > > > > > ______________________________________________________________________ > This email has been scanned by the MessageLabs Email Security System. > For more information please visit http://www.messagelabs.com/email > ______________________________________________________________________ Joel B. Sheffield, Ph.D. Biology Department, Temple University 1900 North 12th Street Philadelphia, PA 19122 [hidden email] (215) 204 8839, fax (215) 204 0486 http://astro.temple.edu/~jbs |
In reply to this post by Jacqueline Ross
We have this "affliction" all the time since we are using a digital
camera on a transmission electron microscope and the phosphor (not being smoothly applied) has a bright quadrant. NIH does an excellent job of subracting the background. Basically, you record an image under the same lighting conditions as the specimen (but minus the specimen). This gives the background that you want to subtract. Next record your image under the same lighting conditions (relative brightness). Open both images, go to Process/Math and subtract the background from the image. It does a super job. Now, I must admit that I am not using Image J but I am sure these features are there as well. John > >I have had an enquiry from someone from another Department who has an >artefact affecting their imaging which looks like a diffraction effect >(e.g. Newton rings) when doing transmitted light microscopy. > > > >I had a look at their microscope and camera system and even when Koehler >illumination is set up properly, you still get this effect. The >microscope is a rather old Nikon inverted microscope with a Coolpix >attached. The objectives, (phase contrast), are not infinity corrected >and I'm assuming that there is also a problem with the tube length for >the camera. The rings are there irrespective whether specimens are on >glass or in plastic dishes. I don't think much can be done optically >(correct me if I'm wrong) but I wondered if some image processing could >help. > > > >My ImageJ question is this: > > > >Is it possible to use some kind of image processing (e.g. image >subtraction) to remove this artifact? I have tried a few things out >(e.g. Image calculator, rolling ball) without much success so far. > > > >I have posted 2 images at the following address: >http://www.health.auckland.ac.nz/biru/exptal_images/index.html There are >also links to larger images although the original images are even larger >than these. > > > >Please note that I do know that these images are appalling! I asked one >of the students to send me one image with tissue (glass slide) and one >without at the same focal plane and this is what I got! I don't think >they set up the microscope very well. However, you can see the rings >clearly on both images so I hope they will inspire someone to suggest a >solution. > > > >I would be very pleased to hear of any potential solutions. > > > >Cheers, > > > >Jacqui. > > > >Jacqueline Ross >Biomedical Imaging Research Unit >School of Medical Sciences >Faculty of Medical & Health Sciences >The University of Auckland >Private Bag 92019 >Auckland, NEW ZEALAND > >Tel: 64 9 373 7599 Ext 87438 >Fax: 64 9 373 7484 > >http://www.health.auckland.ac.nz/biru/ ><http://www.health.auckland.ac.nz/biru/> > > -- ############################################################## John J. Bozzola, Ph.D., Director I.M.A.G.E. (Integrated Microscopy & Graphics Expertise) 750 Communications Drive - MC 4402 Southern Illinois University Carbondale, IL 62901 U.S.A. Phone: 618-453-3730 Email: [hidden email] Web: http://www.siu.edu/~image/ ############################################################## |
In reply to this post by Jacqueline Ross
Although not as effective as a lightfield/darkfield background
subtract, shouldn't it be possible to write a plugin to correct for this automatically? I have no idea how to implement the details, but here's my general scheme: Interpret the image as a radial plot, possibly with averaging. This should give an image with strong vertical or horizontal bands. Use a FFT-based system to isolate the background component, then undo the radial step to generate a synthetic lightfield image. I can't say how easy this would be, but it might be a solution for people who don't have the option for a control-image background subtraction. |
In reply to this post by Jacqueline Ross
Dear Jacqui,
I have used the Coolpix 995 and Coolpix4500 on different compound microscopes and was able to suppress the ring artefact to reasonable levels. The severity of the artefact may vary from camera to camera, but it is apparently caused by the spin-cast manufacturing process of Coolpix lenses: http://www.couger.com/microscope/shootout/shootout.html I found that the ring artefact is less obvious if I zoom out with the Coolpix as much as possible. People sometimes make the mistake of zooming in all the way to maximum to gain (useless) magnification - the rings are then pretty visible. So zoom in only enough to remove the vignetting - it depends on your adapter (eyepiece) how much zoom is necessary - the closer is the eyepiece to the front of the coolpix lens, the better - in my adapter the lenses were only about 1 mm apart, so not much zoom was needed. Some of the genuine Nikon adapters had the front element recessed too much and a lot of zoom was needed to remove vignetting. Of course, if you need reproducible magnification, it would be nice to set the zoom to the same numerical value each time the camera is used. There is some free software that can be used for this purpose and also can be used as a remote shutter. I can provide more detail off-list. On my Olympus BH2 microscope the artefact was also less severe with lower magnification objectives than with the oil 100x/1.3. Regards, Stan At 01:40 AM 7/18/2006, you wrote: >Dear All, > > > >I have had an enquiry from someone from another Department who has an >artefact affecting their imaging which looks like a diffraction effect >(e.g. Newton rings) when doing transmitted light microscopy. > > > >I had a look at their microscope and camera system and even when Koehler >illumination is set up properly, you still get this effect. The >microscope is a rather old Nikon inverted microscope with a Coolpix >attached. The objectives, (phase contrast), are not infinity corrected >and I'm assuming that there is also a problem with the tube length for >the camera. The rings are there irrespective whether specimens are on >glass or in plastic dishes. I don't think much can be done optically >(correct me if I'm wrong) but I wondered if some image processing could >help. > > > >My ImageJ question is this: > > > >Is it possible to use some kind of image processing (e.g. image >subtraction) to remove this artifact? I have tried a few things out >(e.g. Image calculator, rolling ball) without much success so far. > > > >I have posted 2 images at the following address: >http://www.health.auckland.ac.nz/biru/exptal_images/index.html There are >also links to larger images although the original images are even larger >than these. > > > >Please note that I do know that these images are appalling! I asked one >of the students to send me one image with tissue (glass slide) and one >without at the same focal plane and this is what I got! I don't think >they set up the microscope very well. However, you can see the rings >clearly on both images so I hope they will inspire someone to suggest a >solution. > > > >I would be very pleased to hear of any potential solutions. > > > >Cheers, > > > >Jacqui. > > > >Jacqueline Ross >Biomedical Imaging Research Unit >School of Medical Sciences >Faculty of Medical & Health Sciences >The University of Auckland >Private Bag 92019 >Auckland, NEW ZEALAND > >Tel: 64 9 373 7599 Ext 87438 >Fax: 64 9 373 7484 > >http://www.health.auckland.ac.nz/biru/ ><http://www.health.auckland.ac.nz/biru/> > > Dr. Stanislav Vitha [hidden email] Microscopy and Imaging Center Texas A&M University BSBW 119 College Station, TX 77843-2257 tel: 979-845-1129 (main desk) tel: 979-845-1607 (direct link) fax: 979-847-8933 |
In reply to this post by Jacqueline Ross
Dear Stan,
Thanks for your comments and advice on how to reduce the artifact. Interestingly, one of the people using the camera had been using the zoom and getting better results so it seems this may be one way around the problem. If you can send me the information on the free software offline, that would be much appreciated. Consistency is important when doing comparisons and saves continually having to recalibrate. Cheers, Jacqui. Jacqueline Ross Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland, NEW ZEALAND Tel: 64 9 373 7599 Ext 87438 Fax: 64 9 373 7484 http://www.health.auckland.ac.nz/biru/ -----Original Message----- From: ImageJ Interest Group [mailto:[hidden email]] On Behalf Of Stanislav Vitha Sent: 20 July 2006 04:04 To: [hidden email] Subject: Re: Optical artefact removal Dear Jacqui, I have used the Coolpix 995 and Coolpix4500 on different compound microscopes and was able to suppress the ring artefact to reasonable levels. The severity of the artefact may vary from camera to camera, but it is apparently caused by the spin-cast manufacturing process of Coolpix lenses: http://www.couger.com/microscope/shootout/shootout.html I found that the ring artefact is less obvious if I zoom out with the Coolpix as much as possible. People sometimes make the mistake of zooming in all the way to maximum to gain (useless) magnification - the rings are then pretty visible. So zoom in only enough to remove the vignetting - it depends on your adapter (eyepiece) how much zoom is necessary - the closer is the eyepiece to the front of the coolpix lens, the better - in my adapter the lenses were only about 1 mm apart, so not much zoom was needed. Some of the genuine Nikon adapters had the front element recessed too much and a lot of zoom was needed to remove vignetting. Of course, if you need reproducible magnification, it would be nice to set the zoom to the same numerical value each time the camera is used. There is some free software that can be used for this purpose and also can be used as a remote shutter. I can provide more detail off-list. On my Olympus BH2 microscope the artefact was also less severe with lower magnification objectives than with the oil 100x/1.3. Regards, Stan At 01:40 AM 7/18/2006, you wrote: >Dear All, > > > >I have had an enquiry from someone from another Department who has an >artefact affecting their imaging which looks like a diffraction effect >(e.g. Newton rings) when doing transmitted light microscopy. > > > >I had a look at their microscope and camera system and even when >illumination is set up properly, you still get this effect. The >microscope is a rather old Nikon inverted microscope with a Coolpix >attached. The objectives, (phase contrast), are not infinity corrected >and I'm assuming that there is also a problem with the tube length for >the camera. The rings are there irrespective whether specimens are on >glass or in plastic dishes. I don't think much can be done optically >(correct me if I'm wrong) but I wondered if some image processing could >help. > > > >My ImageJ question is this: > > > >Is it possible to use some kind of image processing (e.g. image >subtraction) to remove this artifact? I have tried a few things out >(e.g. Image calculator, rolling ball) without much success so far. > > > >I have posted 2 images at the following address: >http://www.health.auckland.ac.nz/biru/exptal_images/index.html There >also links to larger images although the original images are even larger >than these. > > > >Please note that I do know that these images are appalling! I asked one >of the students to send me one image with tissue (glass slide) and one >without at the same focal plane and this is what I got! I don't think >they set up the microscope very well. However, you can see the rings >clearly on both images so I hope they will inspire someone to suggest a >solution. > > > >I would be very pleased to hear of any potential solutions. > > > >Cheers, > > > >Jacqui. > > > >Jacqueline Ross >Biomedical Imaging Research Unit >School of Medical Sciences >Faculty of Medical & Health Sciences >The University of Auckland >Private Bag 92019 >Auckland, NEW ZEALAND > >Tel: 64 9 373 7599 Ext 87438 >Fax: 64 9 373 7484 > >http://www.health.auckland.ac.nz/biru/ ><http://www.health.auckland.ac.nz/biru/> > > Dr. Stanislav Vitha [hidden email] Microscopy and Imaging Center Texas A&M University BSBW 119 College Station, TX 77843-2257 tel: 979-845-1129 (main desk) tel: 979-845-1607 (direct link) fax: 979-847-8933 |
Free forum by Nabble | Edit this page |