Hi Dan,

On Mon, 17 Mar 2014, Daniel White wrote:

> >From:    Johannes Schindelin <[hidden email]>
>
> >On Sun, 16 Mar 2014, Jeremy Adler wrote:
> >
> >> So I suggest sticking with the Pearson and rank Spearman for
> >> correlation analysis, they have a long history. If you really want to
> >> go binary, try Kendall's tau.
>
> >For your interest: I am happy to report that Coloc 2 recently learnt to
> >calculate Kendall's Tau.
>
> I think I'm worried that most of our statistical assumptions are actually
> false in the case of using Pearsons etc. in coloc analysis of fluorescence
> images of biological systems.... why?
>
> Because the intensities distribution of the fluorescence signal are almost
> never anything close to being normally distributed: a Gaussian,  bell
> curve.

It is true that already a Poisson distribution is a much better model for
fluorescence (because there is no negative fluorescence).

However, we cannot in general assume a Poisson distribution. There are
many challenges in typical images, such as offsets, noise, sometimes
signals which are non-linearly related to the underlying value we want to
measure.

That is why I implemented Kendall's Tau. It only assumes ranked values.
The only remaining problem with that is noise: Kendall's Tau is a
frequentist method that assumes that your data are exact, without flaw. I
guess it would not be too difficult to add a "grace difference" in which
values are considered equal as far as the Tau calculation is concerned
when the numerical values acutally differ. But that looks too ad-hoc to
me.

Ciao,
Johannes

--
ImageJ mailing list: http://imagej.nih.gov/ij/list.html