Dear Shreya,
Your dicom header is almost empty.
For example there is no "modality" tag, that should say that the
modality is CT. Usually we also find a lot of other informations like
the manifacturer name, the XRay kV used ect ect...
I don't think the problem is in Fiji.
Fiji is showing you all you can get from the file and obviously the
unit don't seems to be UH.
I think you should see the manufacturer documentation of your CT device.
I think there are two possibilities :
1) The unit are not UH and your CT might use another scale, in that
case probably that your manufacturer could give you a conversion rule
to get UH.
2) The original might been transformed by another software and some
tag have been deleted or changed leading to a wrong scale. Are you
sure that this file is the original output file from your CT device ?
Best regards,
Salim