Dear ImageJ folks--
Has anyone come up with a good method for background subtraction in images obtained using fluorescence microscopy? In my hands, the methods that work for brightfield don't work for fluorescence. Thanks! Martin -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 E-mail: [hidden email] |
In my opinion (my main experience is ca2+ signals in living cells) the
best method is to acquire a background image in an area of the coverslip where there are no cells, using it to substract to the series of images. Sometimes this is difficult if the preparation is too "dense", and if you have to move too far from the area of interest the plane of focus could be different. Now I´m removing to the entire image the average fluorescence from an irregular ROI drawn in the "empty" area surrounding the cells. Martin Wessendorf wrote: > Dear ImageJ folks-- > > Has anyone come up with a good method for background subtraction in > images obtained using fluorescence microscopy? In my hands, the > methods that work for brightfield don't work for fluorescence. > > Thanks! > > Martin |
yes, here is what i was going to suggest:
While this is not the best method, a low-tech solution is to acquire 'blank' images of your sample and use this for straight subtraction of the background. If you are doing confocal or multiphoton imaging, your 'blank' can simply be an image z-slice in which there is no relevant pixel data (only noise). If you have several such images, you can get an average and subtract the average from each frame in the z-series. If you are using epifluorescence it may be a bit tricky, but you should be able to get a decent 'blank' image by acquiring an image through the same coverslip/dish (whatever you are using) without any specimen present. Again, take a few, get a mean, and subtract the mean from your real data. This low-tech solution can work surprisingly well depending on the statistics of the background you are trying to subtract. bryan On Tue, 20 Sep 2005, Pedro J Camello wrote: .' In my opinion (my main experience is ca2+ signals in living cells) the .' best method is to acquire a background image in an area of the coverslip .' where there are no cells, using it to substract to the series of images. .' Sometimes this is difficult if the preparation is too "dense", and if .' you have to move too far from the area of interest the plane of focus .' could be different. Now I´m removing to the entire image the average .' fluorescence from an irregular ROI drawn in the "empty" area surrounding .' the cells. .' .' Martin Wessendorf wrote: .' .' > Dear ImageJ folks-- .' > .' > Has anyone come up with a good method for background subtraction in .' > images obtained using fluorescence microscopy? In my hands, the .' > methods that work for brightfield don't work for fluorescence. .' > .' > Thanks! .' > .' > Martin .' .' Division of Biology Broad Center Caltech 114-96 Pasadena, CA, 91125 ph. (626) 395-2140 ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^ The illusion that we are separate from one another is an optical delusion of our consciousness. -Albert Einstein ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^ |
In reply to this post by Pedro J CamelloDr Pedro J Camello
I agree with Pedro here. My only other comment is with some older
cameras, the background is not even so subtracting a single value for the backround is not correct and a 'black-field' subtraction may be the better way. If it's a timecourse you may need to take a measure of background for each timepoint too. Tony Pedro J Camello wrote: > In my opinion (my main experience is ca2+ signals in living cells) the > best method is to acquire a background image in an area of the coverslip > where there are no cells, using it to substract to the series of images. > Sometimes this is difficult if the preparation is too "dense", and if > you have to move too far from the area of interest the plane of focus > could be different. Now I´m removing to the entire image the average > fluorescence from an irregular ROI drawn in the "empty" area surrounding > the cells. > > Martin Wessendorf wrote: > >> Dear ImageJ folks-- >> >> Has anyone come up with a good method for background subtraction in >> images obtained using fluorescence microscopy? In my hands, the >> methods that work for brightfield don't work for fluorescence. >> >> Thanks! >> >> Martin -- Tony Collins, Ph.D. Facility Manager Wright Cell Imaging Facility Toronto Western Research Institute 13-407 McLaughlin Pavilion 399 Bathurst Street Toronto, ON. M5T 2S8 tel. (416) 603 5367 fax: (416) 603 5745 http://www.uhnresearch.ca/wcif This e-mail may contain confidential and/or privileged information for the sole use of the intended recipient. Any review or distribution by anyone other than the person for whom it was originally intended is strictly prohibited. If you have received this e-mail in error, please contact the sender and delete all copies. Opinions, conclusions or other information contained in this e-mail may not be that of the organization. |
In reply to this post by Martin Wessendorf
I would imagine that part of the problem is that you are trying to
remove background signal which is mostly dark. This means that the pixel values are close to zero or so. Image subtraction might cause the values to drop below zero, which wouldn't be helpful. Some times I have found it useful to divide the data image by the background image, using a 16 or 32 bit resultant image, and then readjust. Joel > Dear ImageJ folks-- > > Has anyone come up with a good method for background subtraction in > images obtained using fluorescence microscopy? In my hands, the > methods that work for brightfield don't work for fluorescence. > > Thanks! > > Martin > -- > Martin Wessendorf, Ph.D. office: (612) 626-0145 > Assoc Prof, Dept Neuroscience lab: (612) 624-2991 > University of Minnesota Preferred FAX: (612) 624-8118 > 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 > Minneapolis, MN 55455 E-mail: [hidden email] Joel B. Sheffield, Ph.D. Biology Department, Temple University 1900 North 12th Street Philadelphia, PA 19122 [hidden email] (215) 204 8839, fax (215) 204 0486 http://astro.temple.edu/~jbs |
In reply to this post by Martin Wessendorf
Dear ImageJ-ers--
I realize that I didn't articulate the problem particularly well. --We want to make some montages of images. Our illumination is not perfectly even and so parts of the field are illuminated more brightly than others. Thus when we try to make the montage, there are abrupt inconsistencies in the lighting across it. I had thought that the following *should* work: 1) Take an image of a uniform fluorescent field (image A) 2) Take an image of the specimen (image B) 3) Divide B by A; multiply by the mean intensity of A. (--I think that this is what Wayne's plugin does.) However, when we use that method, the inconsistencies of the illumination are greatly accentuated rather than being reduced. My guess is that this is due to non-linearity of response in our CCD camera. I'd be delighted if anyone has a work-around for this problem! Martin -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 E-mail: [hidden email] |
In reply to this post by Martin Wessendorf
Have you tried running the Background Subtractor on each separate image
before making the montage? Although I think there are some minor bugs in it (the bottom few rows show artifacts on contrast stretching) it works quite well at flattening out illumination differences. Jordan Bevic QuadTech Advanced Vision Systems -----Original Message----- From: ImageJ Interest Group [mailto:[hidden email]] On Behalf Of Martin Wessendorf Sent: Tuesday, September 20, 2005 1:12 PM To: [hidden email] Subject: Re: background removal in fluorescence Dear ImageJ-ers-- I realize that I didn't articulate the problem particularly well. --We want to make some montages of images. Our illumination is not perfectly even and so parts of the field are illuminated more brightly than others. Thus when we try to make the montage, there are abrupt inconsistencies in the lighting across it. I had thought that the following *should* work: 1) Take an image of a uniform fluorescent field (image A) 2) Take an image of the specimen (image B) 3) Divide B by A; multiply by the mean intensity of A. (--I think that this is what Wayne's plugin does.) However, when we use that method, the inconsistencies of the illumination are greatly accentuated rather than being reduced. My guess is that this is due to non-linearity of response in our CCD camera. I'd be delighted if anyone has a work-around for this problem! Martin -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 E-mail: [hidden email] |
In reply to this post by Martin Wessendorf
Dear Martin,
Actually, what the background subtraction does is to use the "rolling ball" Image is sampled in a ratio and then degraded by a Structuring Element. then the degraded image is subtracted from the initial image (i.e. "top hat" transform). Anyway, I think you're on the right track. May be you should do stack averaging? Or may be you should track the intensity of a praticular structure and readjust the histograms according to it. best regards Dimiter Date: Tue, 20 Sep 2005 13:11:52 -0500 From: Martin Wessendorf <[hidden email]> Subject: Re: background removal in fluorescence Dear ImageJ-ers-- I realize that I didn't articulate the problem particularly well. --We want to make some montages of images. Our illumination is not perfectly even and so parts of the field are illuminated more brightly than others. Thus when we try to make the montage, there are abrupt inconsistencies in the lighting across it. I had thought that the following *should* work: 1) Take an image of a uniform fluorescent field (image A) 2) Take an image of the specimen (image B) 3) Divide B by A; multiply by the mean intensity of A. (--I think that this is what Wayne's plugin does.) However, when we use that method, the inconsistencies of the illumination are greatly accentuated rather than being reduced. My guess is that this is due to non-linearity of response in our CCD camera. I'd be delighted if anyone has a work-around for this problem! Martin |
Wouldn't you want to just subtract the background image from the target
image? 'B - A' in your example below? Ryan Deaton Data Management Specialist Preventive Medicine Northwestern University ph: 312-503-1979 fax: 312-908-9588 -----Original Message----- From: ImageJ Interest Group [mailto:[hidden email]] On Behalf Of Prodanov, D. (FYS) Sent: Wednesday, September 21, 2005 3:58 AM To: [hidden email] Subject: Re: background removal in fluorescence Dear Martin, Actually, what the background subtraction does is to use the "rolling ball" Image is sampled in a ratio and then degraded by a Structuring Element. then the degraded image is subtracted from the initial image (i.e. "top hat" transform). Anyway, I think you're on the right track. May be you should do stack averaging? Or may be you should track the intensity of a praticular structure and readjust the histograms according to it. best regards Dimiter Date: Tue, 20 Sep 2005 13:11:52 -0500 From: Martin Wessendorf <[hidden email]> Subject: Re: background removal in fluorescence Dear ImageJ-ers-- I realize that I didn't articulate the problem particularly well. --We want to make some montages of images. Our illumination is not perfectly even and so parts of the field are illuminated more brightly than others. Thus when we try to make the montage, there are abrupt inconsistencies in the lighting across it. I had thought that the following *should* work: 1) Take an image of a uniform fluorescent field (image A) 2) Take an image of the specimen (image B) 3) Divide B by A; multiply by the mean intensity of A. (--I think that this is what Wayne's plugin does.) However, when we use that method, the inconsistencies of the illumination are greatly accentuated rather than being reduced. My guess is that this is due to non-linearity of response in our CCD camera. I'd be delighted if anyone has a work-around for this problem! Martin |
Free forum by Nabble | Edit this page |